Jump to content

Sphere qualification


---
 Share

Recommended Posts

I have 4-sided star with really short arms which cause it to shank out during qualification. Is there any way to prevent the tactile scanning from going to the equator so that it does not shank out?

image.jpeg.53bccd4955ca7834755335ad7bab64b1.jpeg

Link to comment
Share on other sites

You can define how you will calibrate your probe - pasive, tensor, dynamic tensor.

I also found you posted a question in another post. That is not working?

Link to comment
Share on other sites

Well we would like to just stick to using the 30mm sphere that we use for all our styli qualifications just to keep it simple. We only have two modes to use for qualification and if we use the geometry re-qualification it actually does use the sphere coverage correctly and will not go to the equator but at the same time it only probes 6 single points and I do not feel that this is a good enough process to qualify the sphere. Scanning would be better but when we use that method it always goes to the equator and ignores the sphere coverage.

Qualification mode.png

Link to comment
Share on other sites

probes 6 single points and I do not feel that this is a good enough process to qualify the "Styli"

Link to comment
Share on other sites

That "Standard" option - what's there?

I am wondering if you can do manual path - i remember when i was doing qualification for disk i had to do manual path - defining own points. Perhaps this would be the way.

But if you have smaller sphere, then it would be better to use that.

Link to comment
Share on other sites

I was told that using probes smaller than ~10% of the diameter of the calibration sphere is not a good idea, but cannot find any official documentation of this. (We use an 8mm sphere for all our smaller probes and it seems to work well) 

Link to comment
Share on other sites

I don't know that I heard the 10% rule (not to mention, that would mean qualifying a 3 mm on a 30 mm reference sphere wouldn't be good and I've never had issues with that). My general rule of thumb is anything 1 mm or below (sometimes 1.5 and below) should be qualified on a smaller reference sphere (usually the 8 mm I think).

 

The sphere coverage doesn't affect the scanning part of passive qualification (it does on the point taking part and the geometry requal), so that isn't really a solution I think. I would think the only real answers is either a longer stem on the stylus or a smaller reference sphere.

  • Like! 1
Link to comment
Share on other sites

  • 1 month later...

For the "passive stylus qualification" CALYPSO will always use 180°, no matter what you type in. For the geometry re-qualification lowering the angle below 180° does work.

 

Best Regards

Marcel

Link to comment
Share on other sites

You will need a smaller reference sphere. As Marcel mentioned above, for a passive sensor, you have to run qual passive at least once. To do this, you have to have a smaller sphere. 

 