[Pa...] Posted 5 hours ago Share Posted 5 hours ago Accura and Accura2, VAST Gold in each My company requires qualification of all styli after drives off, whether it happened due to a collision or a holiday weekend shutdown. I've felt for a long time that this is unnecessary, particularly in the intentional shutdown scenario. What would change about a stylus or a stylus system (relative to the Master Stylus) during a shutdown? I've got a couple machines; each has a rack with 21 systems. They're all very different from one another, meaning I need to use 3 different sphere sizes in a total of 7 positions. A full qualification of both machines takes just under 3 hours. This past holiday the pain of lost hours was great enough I think I'm poised to make a change here. I'd happily verify a couple and get back to work in 15 minutes if there were no findings. Google's AI bot says there's no reason. I believe him, but I don't trust him 😉 What do you guys think? Link to comment Share on other sites More sharing options...
[Ma...] Posted 4 hours ago Share Posted 4 hours ago I would calibrate right after crash just that probe system and then regularly, depending on your usage. You can get into situation, where you have a collision which is not a collision, just a probe which ran into a void. There is no need to do such calibration after drives are off - they can shut down to conserve air consumption every 15 minutes of inactivity - that would be hilarious to do calibration in this case. I am running calibration program, which measure before calibration and then after calibration - this can tell you if you have a problem with specific probe ( lose screw and so on ) Link to comment Share on other sites More sharing options...