[Pa...] Posted 22 hours ago Share Posted 22 hours ago Accura and Accura2, VAST Gold in each My company requires qualification of all styli after drives off, whether it happened due to a collision or a holiday weekend shutdown. I've felt for a long time that this is unnecessary, particularly in the intentional shutdown scenario. What would change about a stylus or a stylus system (relative to the Master Stylus) during a shutdown? I've got a couple machines; each has a rack with 21 systems. They're all very different from one another, meaning I need to use 3 different sphere sizes in a total of 7 positions. A full qualification of both machines takes just under 3 hours. This past holiday the pain of lost hours was great enough I think I'm poised to make a change here. I'd happily verify a couple and get back to work in 15 minutes if there were no findings. Google's AI bot says there's no reason. I believe him, but I don't trust him 😉 What do you guys think? Link to comment Share on other sites More sharing options...
[Ma...] Posted 22 hours ago Share Posted 22 hours ago I would calibrate right after crash just that probe system and then regularly, depending on your usage. You can get into situation, where you have a collision which is not a collision, just a probe which ran into a void. There is no need to do such calibration after drives are off - they can shut down to conserve air consumption every 15 minutes of inactivity - that would be hilarious to do calibration in this case. I am running calibration program, which measure before calibration and then after calibration - this can tell you if you have a problem with specific probe ( lose screw and so on ) Link to comment Share on other sites More sharing options...
[Da...] Posted 11 hours ago Share Posted 11 hours ago (edited) I'm using quite a number of different probe systems. If I did a complete re-qualification every day, I wouldn't do much else. So, I have to be smart about it. I want to make sure of two things: The probe systems measures the correct diameter. Instead of a whole calibration routine, I run a small routine that measures the reference sphere. That way I can make sure the measured diameter is within limits. If I use multiple probes in one program, I re-calibrate those (and only those) ahead of my first run of the day. Usually, I check the probes more often than I calibrate them. It's faster and gives you a feeling of security, because I had calibrations that I had to repeat, because the positions among the probe systems were not satisfyingly precise. Edited 11 hours ago Link to comment Share on other sites More sharing options...
[Pa...] Posted 3 hours ago Author Share Posted 3 hours ago So far, just as I thought, except I didn't know about the drives shut down after 15 minutes. That's a new one on me. Is there a way I can validate this and show it to the people requiring this needless exercise? Yes, qualification after collision is certainly necessary. Link to comment Share on other sites More sharing options...
[No...] Posted 3 hours ago Share Posted 3 hours ago They only shut down if the CMM is equipped with AirSaver. Then you have a small application on the PC where you can set the timeout and also shut them off manually. The default is 15 minutes. This thread has a screenshot of the app: https://qualityforum.zeiss.com/topic/18466-air-saver-help/ 1 Link to comment Share on other sites More sharing options...
[Je...] Posted 1 hour ago Share Posted 1 hour ago I would recommend you ask for proof of the claim that verification is needed after CMM shutdown. In most situations, it is not risk-mitigating or necessary to requalify all styli after a power shutdown. Link to comment Share on other sites More sharing options...