[Ju...] Posted January 21 Share Posted January 21 Hello, Recently purchased a new Spectrum with Calypso 2025 after using 2019 for years. An issue I am seeing is with the "Define limit value for Stylus", not alerting me when my styli have gone longer than the set hours for calibration. So, I have it set to 12 hours, but the probe will still measure after that time. I have tried setting this globally in Settings, and the probe individually and the issue still exists. Any help would be greatly appreciated! Link to comment https://qualityforum.zeiss.com/topic/25621-stylus-limit-values-not-functioning/ Share on other sites More sharing options...
[Ke...] Posted January 22 Share Posted January 22 What is the behavior set to in "Extras-Settings-CMM-Stylus System-Stylus check during CNC start"? 1 Link to comment https://qualityforum.zeiss.com/topic/25621-stylus-limit-values-not-functioning/#findComment-144993 Share on other sites More sharing options...
[Pi...] Posted January 22 Share Posted January 22 Please sign in to view this quote. I dont see on this picture clicked tick boxes to active specific values - all of them are turned off. Please activated them and then make test. 1 Link to comment https://qualityforum.zeiss.com/topic/25621-stylus-limit-values-not-functioning/#findComment-144999 Share on other sites More sharing options...
[Ju...] Posted January 22 Author Share Posted January 22 Thank you Piotr and Kevin. I believe your recommendations have solved my issue! 1 Link to comment https://qualityforum.zeiss.com/topic/25621-stylus-limit-values-not-functioning/#findComment-145005 Share on other sites More sharing options...