[Ju...] Posted Wednesday at 08:16 PM Share Posted Wednesday at 08:16 PM Hello, Recently purchased a new Spectrum with Calypso 2025 after using 2019 for years. An issue I am seeing is with the "Define limit value for Stylus", not alerting me when my styli have gone longer than the set hours for calibration. So, I have it set to 12 hours, but the probe will still measure after that time. I have tried setting this globally in Settings, and the probe individually and the issue still exists. Any help would be greatly appreciated! Link to comment Share on other sites More sharing options...
[Ke...] Posted 12 hours ago Share Posted 12 hours ago What is the behavior set to in "Extras-Settings-CMM-Stylus System-Stylus check during CNC start"? 1 Link to comment Share on other sites More sharing options...
[Pi...] Posted 12 hours ago Share Posted 12 hours ago Please sign in to view this quote. I dont see on this picture clicked tick boxes to active specific values - all of them are turned off. Please activated them and then make test. 1 Link to comment Share on other sites More sharing options...
[Ju...] Posted 10 hours ago Author Share Posted 10 hours ago Thank you Piotr and Kevin. I believe your recommendations have solved my issue! 1 Link to comment Share on other sites More sharing options...