[Ch...] Posted Tuesday at 08:06 PM Share Posted Tuesday at 08:06 PM I'm using software GOM Professional V8 on a GOM Atos 3 rev 2. I just got this machine from an affiliate and we really don't know anything about it. Here's my issue, I'm trying to run the calibration, but the dropdown to import the calibration file is greyed out(see attached). I've checked and the calibration files for this plate are in the appdata folder. But I don't know how to load it. Link to comment Share on other sites More sharing options...
[Ma...] Posted 17 hours ago Share Posted 17 hours ago On our Atos5 you are loading calibration files via web page ( go to setup measuring setup via help / System analysis ) and on ATOS device is a button for web service. Also before calibration you should setup / ensure you have selected correct measuring volume. Then it should use only correct calibration plane with dots. For MV it should be a specific one. Can you click on Edit button? 1 Link to comment Share on other sites More sharing options...
[Na...] Posted 17 hours ago Share Posted 17 hours ago Hello Chad, can you load the files into the preferences -> templates -> calibration objects? Should appear in that dialog afterwards. 1 Link to comment Share on other sites More sharing options...
[Ch...] Posted 16 hours ago Author Share Posted 16 hours ago preferences -> templates -> calibration objects This allowed me to import the correct calibration file. Will try this very soon. I'm pretty confident we have the correct plate for our lenses and measuring volume. The web is an issue for us at the moment. It's a windows 7 computer and our IT department will only allow Windows 11 and Windows 10 devices with extended support to connect to our network or the web. They tend to make lots of special exceptions for me and we're working on a solution. I'd like to upgrade to Zeiss Inspect using a new version of windows, but that a future project. I feel like I'll run into all sorts of compatibility issues once I try to upgrade the computer and software. Link to comment Share on other sites More sharing options...
[Ma...] Posted 14 hours ago Share Posted 14 hours ago Web part is only connection to ATOS device ( it's IP ) so it should'n trigger anything. I saw webserver is running at localhost:PORT and for device control you would have IP for device ( not sure how ATOS 3 is with this ) Our ATOS 5 is connected to PC via 10GBps optical lan card - how is that on ATOS 3? Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in