Problems calibrating probe


---
 Share

Recommended Posts

 
 
 
 
 
327
 
Good morning, I've purchased a probe with a 0.3 mm diameter (extremely small), and when I calibrate it, the result is insufficient. I think it's because during the calibration process, the probe catches on the probe shaft. Where can I change the settings so that it calibrates successfully? Thank you very much.
 
Link to comment
Share on other sites

Please sign in to view this quote.

Hello again, I'm still having problems and it's still giving me insufficient calibration. I've even lowered the degrees to 90, but when the dial makes the first circular sweep, it still hits the stem... I don't know if that's the problem. I don't know if I should also increase the limit values, the sigma, and so on...

Link to comment
Share on other sites

What probehead are you utilizing? If it is an xxt system, changing the angular coverage will not have the desired effect, your only option would be to acquire a smaller aphere so that the probe does not hit. 

Link to comment
Share on other sites