Jump to content

CMM Reports and Rounding...


---
 Share

Recommended Posts

This is my first post in the Zeiss forum so bear with me if this has already been answered.

 

I have been asked to look into reformatting the report generated from the CMM to see if there was a way to change how CALYPSO makes its judgement on the report.

Our report currently generates measurements out to the 6th decimal place (.XXXXXX) and makes a red/green judgement accordingly.

 

As much as I disagree, the task is to reduce the measured value to one decimal place past the part tolerance and judge based off this value.

I.e., if a part specification is 2.000 +/- .010 and the part measures 2.010040, the wish is for the CMM report to show 2.0100 and a green judgement. 

 

As I stated, I do not agree with this methodology because it is not in accordance with Section 5.4 of ASME Y14.5-2018.

 

I'm also learning how CALYPSO and PiWeb tie together so I'm not sure if this is something that can easily be done in PiWeb or vise-versa.

 

I appreciate any feedback to this.

 

 

  • Like! 1
Link to comment
Share on other sites

Hi

Please sign in to view this username.

.

Thanks for creating your first post, and I look forward to reading many more from you.

I think the scenario you presented about rounding and significant digits is one that many of us have encountered.  So, yes, it has been discussed on these forums, but by no means is it "answered."  It's an ongoing discussion.

Luke's reference to a similar post is helpful.

My stance:
------------

The associates who are encouraging you to show "green" on the report when a numerical size limit has been exceeded (which is effectively what they are asking for) should reconsider their request.

They are not asking for truncation to significant digits.  They are requesting you to round.  True metrology software should guard against this behavior, and I'm grateful that Calypso does exactly that.

In any situation in my professional career when this has been requested, I have informed the person making the request that they are welcome to make the proposed alterations with their own ink and signature.  I will not.

  • Like! 3
  • Awesome! 1
Link to comment
Share on other sites

Thank you all for the feedback and this is even more ammunition I need to help push back.

My inspector and I dug into CALYPSO earlier this morning investigating this and we both agreed that doing this is completely not a good idea.

It goes against the very standard that our drawings are built around. 

 

The Y14.5 is black and white, plain and simple. 

All values are absolute, regardless of how many decimal points are reported and measured.

  • Like! 2
Link to comment
Share on other sites

I agree with the gentlemen above, the only potential argument against this is :

What is the 'actual' uncertainty of the machine, which is a long an drawn out process to truly calculate, in which case sure there could be some wiggle room - however it also could potentially be worse than reported - which then perhaps guard banding may be required, etc.

The other great cases for this situation are (😉) :

a) What is .0001 between friends ? 😁🤣

b) .0001 - that will be gone as soon as they fire up the chopper 😁😂

Welcome to the world of measuring. 

Link to comment
Share on other sites

About that wiggle room though - Measurement uncertainty cannot ever be honestly or appropriately used to accept product rejected by the measurement process.  It can only drive you to reevaluate using a process with a smaller calculated uncertainty, or to accept the uncertainty (and thereby the failures) you have.

We have a series of tools here; basically hard gages of known lengths used as a comparative measurement to validate 100"+ lengths.  When I first arrived, uncertainty was the buzz word around here.  All of these tools were re-categorized and relabeled.

What had once been a hard gage with a length of 186.1324" now became a block labeled 186.132 ± .0013.  I guess the intent was to communicate potential error to the machinists so they could use "common sense" (another word for "unicorns") to accept product.  Accept they did.

Suddenly nobody ever cut anything out of tolerance for length, but the CMM at final inspect at THE SAME MOMENT suddenly became absolutely incapable of sentencing any ± .03" lengths. We continued to use it to validate 10.000 ± .00015" diameters though, as it apparently didn't have trouble with those.

Must've been those unicorns.

Just ask yourself:  Who's risk am I trying to mitigate here?  If it's a manager hellbent on one more shipment for the month, and he's asking you to transfer the risk of a short month onto the women and children on those airliners my parts go into as potential risk to life, I guess I'd need to find a different CMM to run.

Hope it ain't a Hexagon...

  • Like! 3
Link to comment
Share on other sites

I agree with

Please sign in to view this username.



Measurement certainty cannot be used to extend a print-specified tolerance range.

  • Like! 1
Link to comment
Share on other sites

  • 4 weeks later...
---

Little late to the party and love the discussion. For those interested in making CALYPSO turn green for the rounded off decimals here is a link on how to do it.

It is a dirty trick. And again as discussed not really suggested but some users understand this and allow for this exception to be used in their process. 

https://portal.zeiss.com/knowledge-base?id=2541215

 

Link to comment
Share on other sites

---

Sometimes our job is the same everywhere and also entertaining. We recently discussed this with the head of the CNC department and switched to 3 decimal places, but then the plant manager said that wasn't possible and we had to go back to 4 decimal places. Of course, CNC often has a “green part.”

 

But the customer is looking for microns, so I would say min. 4 decimal places is standard.

 

Sorry, I'm writing from Germany, so 0.001mm (metric), 0,00003937008

 

It's always a battle with windmills.👻

 

 

Link to comment
Share on other sites

 Share

×
×
  • Create New...