My colleague is recommending that the operator remove the probe from the probe head prior to machine shutdown. Is there an issue with leaving the stylus system in the CMM head while the machine is off? Or is it best practice to remove the current probe from the probe head prior to shutdown?

I also commented to not overcomplicate the operator's instructions as there could be a risk of messing something up with them using the automatic changer.

I didn't think this was an issue, but I wanted to double-check.

TIA!

---

In our setup we swap the styli in the rack with almost every part. Therefore the Autorun-only operators needed a way to remove all the styli from the CMM without having to mess around with the joystick box or the like. Our foolproof solution is to have an empty probe plate in a special place in the rack at all times, which can be fetched by a special program when a part change is due or the machine has to be shut down.

A possible issue with an arbitrary stylus being left in the head is that it may collide with the rack when the machine is homed. But I remember that some years ago they added a Calypso function where you can mark each stylus as to whether it is "homing-proof" or not. But I can't remember how to get there.

Leaving the CMM without a stylus in the head carries the risk of particles accumulating at the magnet.

  • Like! 1
---

Please sign in to view this quote.

Please sign in to view this username.

 This is correct, I have ours set to not home with any stylus in the head. It throws a warning if there is a stylus present, and then lets you drop it before homing. Off the top of my head I can't recall the location for the setting, but probably not too hard to find by poking around in the CMM settings. 

  • Like! 1
---

Removing probe from head would be easier on XXT, but there are usually not so heavy probes.

Vast on the other hand can have benefits for removing probes, but i was told that machine is so much whining about no probe, that we just keep last probe there.
None of the probes are risk for collision while homing.

---

The only time I remove a probe from the sensor is when I have minimal clearance or no clearance to a top tier probe rack such as I have on 500,500,500 Micura's.

I don't have the room for having the master probe in the sensor if a machine home is required.

Other than that, I do not remove the probe from the sensor.

  • Like! 1
---

Absolutely no issue leaving your stylus systems in the head when shutting down. In fact it is actually recommended to leave something in the sensor to keep the contacts from being exposed. 

Only thing to consider is homing issues, if you leave a very awkward styli (long/wide) in the sensor and home the sensor after the reboot, you may have a collision with homing navigation. Which a few posters above have given awesome recommendations to prevent this. 

  • Like! 2
  • 2 weeks later...
---

My CMMs are never stored with an empty sensor.  I wouldn't allow it.  One of them has not been stored empty in over 20 years.  This guidance is misguided.

VAST Gold.  I have no problem leaving a system 60" OAL in the sensor over a long weekend.

That said, we only turn our machines off 3 times per year at most so...

I'd say that's also optional.

---

I'm going to give away my age here, but that's ok. Where the practice of removing styli from heads came from was way back with the first CMMs. Many were totally manual where the operator had to physically move the head around so the probe could make contact with the part. Some were hard probe, some were touch probe. There weren't any motors or joysticks, 100% manual. All 3 axis' would glide very freely on a cushion of air. The only thing an operator could do is lock or unlock each axis. What would happen then is at night when everybody was going home, the air compressors would get shut down and the air pressure in the lines would drop. If the Z axis in particular wasn't locked in position, the Z mast would drop and smash the stylus system. As CMMs became automated, the problem of the Z mast dropping went away, but the practice of removing stylus systems at night kept on for a long time. Now it's simply a matter of choice, but you still need to be aware when homing as others have pointed out.

  • Like! 3
×
×
  • Create New...