My colleague is recommending that the operator remove the probe from the probe head prior to machine shutdown. Is there an issue with leaving the stylus system in the CMM head while the machine is off? Or is it best practice to remove the current probe from the probe head prior to shutdown?

I also commented to not overcomplicate the operator's instructions as there could be a risk of messing something up with them using the automatic changer.

I didn't think this was an issue, but I wanted to double-check.

TIA!

Link to comment
Share on other sites

---

Never have I ever - removed the probe from the probe head prior to machine shutdown... at least in the last 30 years 😂

If anything, I prefer to have the Master Probe loaded prior to an intended shutdown of the machine.

  • Like! 3
Link to comment
Share on other sites

---

In our setup we swap the styli in the rack with almost every part. Therefore the Autorun-only operators needed a way to remove all the styli from the CMM without having to mess around with the joystick box or the like. Our foolproof solution is to have an empty probe plate in a special place in the rack at all times, which can be fetched by a special program when a part change is due or the machine has to be shut down.

A possible issue with an arbitrary stylus being left in the head is that it may collide with the rack when the machine is homed. But I remember that some years ago they added a Calypso function where you can mark each stylus as to whether it is "homing-proof" or not. But I can't remember how to get there.

Leaving the CMM without a stylus in the head carries the risk of particles accumulating at the magnet.

  • Like! 1
Link to comment
Share on other sites

---

Yeah, I would typically leave the prob on during shutdown, I have never had any issues doing this. 

Thanks for the feedback guys!

Link to comment
Share on other sites

---

Please sign in to view this quote.

Please sign in to view this username.

 This is correct, I have ours set to not home with any stylus in the head. It throws a warning if there is a stylus present, and then lets you drop it before homing. Off the top of my head I can't recall the location for the setting, but probably not too hard to find by poking around in the CMM settings. 

  • Like! 1
Link to comment
Share on other sites

---

Please sign in to view this quote.

This would be Settings > CMM > Probe > Check stylus.

  • Like! 3
Link to comment
Share on other sites

---

Removing probe from head would be easier on XXT, but there are usually not so heavy probes.

Vast on the other hand can have benefits for removing probes, but i was told that machine is so much whining about no probe, that we just keep last probe there.
None of the probes are risk for collision while homing.

Link to comment
Share on other sites

---

Thanks for the feedback all!

When introducing new probes to the tree for long-term use, I typically verify that there will be no collisions when homing. 

I learned something new on the check stylus setting! Thanks again guys!

Link to comment
Share on other sites

---

The only time I remove a probe from the sensor is when I have minimal clearance or no clearance to a top tier probe rack such as I have on 500,500,500 Micura's.

I don't have the room for having the master probe in the sensor if a machine home is required.

Other than that, I do not remove the probe from the sensor.

  • Like! 1
Link to comment
Share on other sites

---

Absolutely no issue leaving your stylus systems in the head when shutting down. In fact it is actually recommended to leave something in the sensor to keep the contacts from being exposed. 

Only thing to consider is homing issues, if you leave a very awkward styli (long/wide) in the sensor and home the sensor after the reboot, you may have a collision with homing navigation. Which a few posters above have given awesome recommendations to prevent this. 

  • Like! 1
Link to comment
Share on other sites

  • 2 weeks later...

My CMMs are never stored with an empty sensor.  I wouldn't allow it.  One of them has not been stored empty in over 20 years.  This guidance is misguided.

VAST Gold.  I have no problem leaving a system 60" OAL in the sensor over a long weekend.

That said, we only turn our machines off 3 times per year at most so...

I'd say that's also optional.

Link to comment