[Jo...] Posted Friday at 07:02 PM Share Posted Friday at 07:02 PM I have 3 angles that won't calculate and are purpled. When I go to the feature of the plane, the strategy and evaluation are correct, and I can run the feature and the probe traces it normally, but it will not resolve the issue. When I click the polyline path of the plane feature it is showing points taken as normal. Link to comment Share on other sites More sharing options...
[Ri...] Posted Friday at 09:47 PM Share Posted Friday at 09:47 PM Please show a screenshot of the planes from the main geometry side as well as the strategies, and the evaluation side. 1 Link to comment Share on other sites More sharing options...
[Jo...] Posted 7 hours ago Author Share Posted 7 hours ago 1 Link to comment Share on other sites More sharing options...
[Jo...] Posted 7 hours ago Author Share Posted 7 hours ago Above are the Screenshots you requested 1 Link to comment Share on other sites More sharing options...
[DW...] Posted 7 hours ago Share Posted 7 hours ago Please sign in to view this username. This seems....high. Change it to 5 and see if it calculates. 2 Link to comment Share on other sites More sharing options...
[Ri...] Posted 6 hours ago Share Posted 6 hours ago Ding ding ding. You outlier eliminated all of the points. lol. 1 Link to comment Share on other sites More sharing options...
[Jo...] Posted 6 hours ago Author Share Posted 6 hours ago Thank you, the typo of 50 instead of 5 was indeed the issue. Thankfully it is only on that one feature. 2 Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment