Jump to content

Honeycomb measurments.


---
 Share

Recommended Posts

Hello GOM team,

I am currently working on a correlation study between a Scancobot with Atos Q and a CMM using Calypso. We are measuring a radius section of 194 mm on a honeycomb part.

We have observed differences in the measurements obtained from both pieces of equipment. Specifically, we are measuring three sections at different heights, over an approximate length of 15 mm, with 7 points per section.

On the Scancobot, we use a touch point edge measurement with an edge height of 4 mm. On the CMM, we measure the same points but use a 2.5 mm radius probe.

We have noticed that the point-to-point variation is greater on the Scancobot than on the CMM. The dispersion between points on the CMM is about 0.01–0.03 mm, whereas on the Scancobot it is 0.04–0.07 mm.

Do you have any recommendations for measuring this type of part?

image.png.2f3b5d80b6ff743713d5285078a80a27.png

image.thumb.png.0786d27c477aa977f1f0ab5f15b9697a.png

Link to comment
Share on other sites

Put some green highlighted points(or section cut) you are using to create a radius on that mesh....no one knows where you are measuring the radius at with just a picture of a mesh.

RADII.jpg