[Dr...] Posted Monday at 07:52 PM Share Posted Monday at 07:52 PM Hello, Currently having issues trying to qualify our rotary table. It is portable and has been moved, therefore we need to qualify to ensure accuracy. Hoping someone would be able to give some guidance on how to qualify or send me in the direction of instructions that are better than the manual given with the rotary itself. Link to comment Share on other sites More sharing options...
[Ca...] Posted 4 hours ago Share Posted 4 hours ago Hi Drew! Are you looking to find the center of the RT again since it moved? If so, you can set the center by opening an empty measurement plan then use the MasterProbe to create a circle around the RT. Under Resources - Rotary Table - Rotary Table Position you can then select the circle for the center feature. Then, you can either run the axis program or start running your parts if you measure the RT axis in the programs. The axis program is typically found here, if needed, C:\Users\Public\Documents\Zeiss\CALYPSO 8.0\workarea\inspections\inspections_rt_axis\RT-Axis. Link to comment Share on other sites More sharing options...