"GD&T Flatness is very straight forward. It is a common symbol that references how flat a surface is regardless of any other datums or features."

This has always been my understanding of flatness, very straight forward, and only checked to itself, low to high points.

So can someone help me understand why the alignment of the part has an affect on flatness?

I wanted to just take a quick look at a surface, so I made a plane and a circle to locate the part. Rotation in space = plane, Z origin = plane, X/Y origin a simple circle in the bore, the part in rotationally symmetric so no Planar rotation. My flatness result is .0017"

So I changed the alignment, added an additional circle in the bore, created a 3d line using the 2 circles. Rotation in space = 3d line, Z origin = plane, X/Y origin = 3d line, no planar. Result is .0007", which is much closer to what I can see manually running and indicator over the surface on a granite table.

If flatness is only checked to itself, why does my alignment affect the result? Flatness is so simple that I feel really dumb that I can't wrap my head around where the extra .001" is coming from, because clearly I am introducing it.

 

Thanks.

Link to comment
Share on other sites

---

Are you measuring the exact same area of the plane each time with the same measurement strategy? Are you taking enough points to adequately define the plane? 

BTW... if you are putting the part on a granite plate and running an indicator across it, you're checking parallelism, not flatness. 

Link to comment
Share on other sites

Please sign in to view this quote.

It's the same plane used for the alignment. I only have 1 plane and 2 circles in the very quick and basic program.

Please sign in to view this quote.

Yeah, I do know that, I just figure if the flatness is at .0017", I should see some movement on the indicator even checking on a granite.

I could nearly duplicate and see the .0007" results checking this way however.

 

If I use the plane1 & circle1 alignment, I get .0017". If I use The 3d line1 and plane1, I get .0007". Same plane is being measured and used for the alignment.

Untitled01.png

Untitled02.png

Untitled03.png

Link to comment
Share on other sites

Was that plane scanned twice? Or how you tested your theory about alignments modifying flatness results?

Is it replicable - aka you run 1st as plane, then 2nd as 3d-line, 3rd as plane and still those results?

Link to comment
Share on other sites

Please sign in to view this quote.

I'm using a different part today, so the results will vary.

Using the bad alignment of Plane1 & Circle1 as pictured before, I take 4 points to grab the plane, then 4 points in the bore for Circle1. Start CNC, then a 1250 point Plane1 is scanned, then a 500 point circle (semi-circle technically to avoid a keyway) is taken. Flatness result is .00198"

I change the alignment to 3d line1 & Plane1, take my 4 point Plane1, then a 4 point Circle1, then a 4 point Circle2 over 2.5" deeper into the bore. Start CNC, 1250 point Plane1 is scanned, then the 500 point Circle1, then another 500 point Circle2. Flatness result is .00066".

Since you asked for it, I now switched BACK to the bad alignment of Plane1 & Circle1, 4 points, 4 points, CNC, result is .00199"

At no point during this has the part been moved, but if I chose to 'clear existing results' and keep current alignment, it will rescan the Plane1 and only Plane1, and the results do not change since it does not retake the new alignment.

 

I hope I explained that clearly enough.

Link to comment
Share on other sites

Thinking about this some more, I think I may know what's going on, and I guess it does make sense, hopefully someone can confirm what I am thinking is true.

The Flatness measurement still needs a perfect form or median value to evaluate from, and when I give the program a 3d line for the rotation in space, it places the plane1 at 90° to the 3d line. But when I use the Plane1 as my rotation is space, it is probably comparing that plane to the perfect form or median plane that would be created by the machine itself. So, my CMM stage has some tilt to it, and that is what I am picking up.

 

Would that be correct?

Link to comment
Share on other sites

I can't replicate this no matter what I do on my offline seat.  This tells me its unlikely its the alignment per say (more on this in a minute).  I don't think your theory about the machine tilt is what's causing it either.  

But here's my theory:   Since changing your spatial datum in the base alignment is clearly causing a repeatable difference, it's possible you are seeing some sort of probe radius correction error.

image.png.07a7f0d729f09ebf7bfab015bce4a79b.png 

 

Is this a rough part, like a casting or plastic?  What kind of difference do you see between using filters vs no filters?  I'm just pontificating here, but maybe Plane 1 is too small to establish a reliable spatial datum; maybe that's why you get better results when you use the larger 3d line in the base alignment. 

 

Anyone else got any ideas?

Link to comment
Share on other sites

If you're manually probing your features are you cleaning up the nominals/vectors of all the features? The software should be creating the planar tolerance zone relative to the feature alignment & feature nominals related to that alignment. Flatness should be best fitting the tolerance zone. Are you measuring the plane as a separate feature not included in your Base Alignment?

Edited
Link to comment
Share on other sites

Please sign in to view this quote.

Ground surface and ground bore (see picture)

Please sign in to view this quote.

about 1 tenth.

Please sign in to view this quote.

It is small, no denying that. It is taken with a circle path on a 2.000" Ø

Please sign in to view this quote.

I don't understand what you are saying, sorry. I included a screenshot of the feature plane1 window. ELI5, Feel free to RESPECTFULLY talk down to me, if that makes sense.

Please sign in to view this quote.

No, as I mentioned before, I have 1 plane and either 1 circle or 2 circles to create a 3d line., check the screenshots I previously posted to see for yourself. This is a very simple basic program that I threw together to simply check a ground surface flatness. This isn't even a print requirement, just me trying to help the shop troubleshoot some issues.

20260511_123525.jpg

Untitled04.png

Link to comment
Share on other sites

Can you make a secondary alignment that is either the 3d line or plane (whichever one is not the current base alignment) and set your feature to it and see if the form changes? It shouldn't. If it doesn't I would see how perpendicular your 3d line is to plane1. I suspect you scanned a different area, maybe you got closer to the keyway, and that's why your number changed. I would also look at the actual points and see where your min and max points are moving to based on the alignment, maybe that will clue you in to what is happening. 

Link to comment
Share on other sites

Please sign in to view this quote.

I think I did what you asked. The form changes. The bore is NOT very perpendicular to the face, this is the very issue I have been trying to help them solve, there is absolutely face runout on this face when compared to the bore, which is translating into the gear when it gets ground, because they are colleting on a small section of the bore, and resting on the ground face. I know the bore and face not being very perp to each other is an issue, but the parts are already done, I am simply trying to help them understand what has happened.

As far as scanning a different area, its the same 2.000" Ø circle path, centered on the alignment/bore.

And the 2 circle paths for the circles in the bore, are both 500 points, start angle of 0°, then I have it set to start at 120° and scan for 300°. Basically I only scan CCW from 11:00 to 1:00, and I simply place the part of the stage with the keyway aligned at 12:00. I have to use a 3.0 mm long probe to take the 3d line, otherwise I'll risk shafting out. And the keyway is very thin and shallow, so I simply am ignoring it and programming around it, rather than using it for a Planar rotation and I would normally do to give a fullly defined alignment.

Untitled05.png

Untitled06.png

Untitled07.png

Untitled08.png

Untitled09.png

Link to comment
Share on other sites

If evaluating flatness is the primary goal, I would stick with using the plane as the spatial axis and a single circle path as the origin. The 3d line looks to be introducing error in the measurement by skewing the axis ever so slightly, thus placing your planar scan path in a slightly different position than the original alignment. 

To test this, would you be able to add a few more scan paths on your plane at different diameters? Maybe one smaller and one larger than your original path? It would be helpful to see the form of the plane over as much as the surface as you can. 

Ensure the proper filters/outliers and eval method (min. feature) as usual and adjust your speeds accordingly.

From there, if still curious, measure the larger plane again against the 3d line alignment and check the flatness plots. How do they compare? Are the highs and lows in relatively the same locations? Are they the same magnitude? 

Link to comment
Share on other sites

The alignment does not change the flatness. But you should consider, that if you use filtering and outlier removal, that there might be a number of points for your plane that borders on "filtering works" or "filtering doesn't work". Sometimes the filtering breaks down if there aren't enough points for the filter to work. And since the filtering parameters can change from element to element and characteristic to characteristic, it might be worth to check every element and characteristic for the right parameters, and also check the default protocol for information about the filtering (it tells you when the filter doesn't work).

Link to comment
Share on other sites

Now I'm just confused even more. So, I was getting frustrated with this whole situation not making sense, so I abandoned the program I had and started over. I was browsing the Zeiss Basic training instructions again, and had the idea to use the a cone to locate the part instead.

I took a cone, plane and symmetry point in the keyway(cone/plane/sym as base align), fully defined alignment as it should be. I even swapped to a different probe to reach into the keyway, again, just to see what changes. I still added the 2 circles and created a 3d line, created 3 additional different alignments, (plane/circle), (Plane/circle/sym), and (3d line/plane) so I could see what changes without going back and forth and re-running the part. Made new planes from recalled points, but changed their alignments. As expected, nothing changes, because the alignment shouldn't matter. I've included a bunch of screenshots of the results and am at least now seeing just over .0001" of deviation between results, and are believable for what a ground surface should measure.

What was happening before still doesn't make sense to me, especially since I recreated the same "bad alignment" program and got good results this time. Only thing I can think of is the fact that the 3.0mm long probe that is being used is a manual load, not from a probe rack, but I'm not getting the 'probe contacts closed' error, and the probes all re-qaul just fine.

Is anyone able to make any sense out of what may have been going wrong previously? I just want to understand what's going on here.

Untitled10.png

Untitled11.png

Untitled12.png

Untitled13.png

Untitled14.png

Untitled15.png

Untitled16.png

Untitled17.png

Untitled18.png

Untitled19.png

Untitled20.png

Link to comment
Share on other sites

The alignment isn't going to change the actuals of the measured plane flatness if you are just doing a recall feature points (they are still the exact same points collected). 

The alignment will absolutely have an impact on the measured plane flatness because each alignment is telling the hardware where to go to measure data. 

It seems like right now you are just throwing a lot of stuff at the wall, and trying to see what sticks, and unfortunately you have muddied the water by introducing so many variables, and not controlling any of them. 

Here's where I would start, and go from there:

  • Create a new measurement plan. 
  • Use a stable base alignment:
    • Use the face on the gear as it is the largest as the Spatial Rotation and Z Origin
      • Only take (4) touch points equally spaced around the face (measure as a Plane)
    • Use the ID of the gear as the X, and Y Origin
      • Only take (4) touch points equally spaced around the diameter (measure as a Circle)
    • Use the Keyway as the Planar Rotation
      • Only take (2) touch points (measure as a symmetry point)
    • Loop the Base Alignment 3x
      • Place a break condition of 0.005mm
  • Measure the Plane on the hub face with (2) circle paths
    • (1) inside and (1) outside
      • Scan at 5mm/sec
      • 0.25mm point spacing
      • Filter set to 2.5mm
      • Outlier Elimination set to ±3 sigma (I wouldn't do adjacent points for a Flatness unless necessary)

The speed/point spacing/filter might need to be adjusted to match your exact size, but I took a stab in the dark based on how it look. 

Link to comment
Share on other sites

Please sign in to view this quote.

I absolutely am, thought that was clear. I am trying to do what everyone is asking me to try, and seeing what changes in an attempt to understand the software better and where my issue/misunderstanding lies. So let me first clear up a little bit, I am NOT a 'CMM programmer', I AM the entire inspection dept. So using the O-inspect 543, is only like 8% of what I do right now. Last time I touched this machine was probably a month ago. I am the first person at this company that has had any understanding of GD&T, before me their handling of GD&T was "We don't know what those symbols mean, so we just ignore them". My mind was BLOWN when I heard that. I took the Zeiss Basic training, but it was at least 6 months AFTER the training, that I was ever able to use the machine. this is not an ideal situation, I WANT to be better at this than I currently am.

Please sign in to view this quote.

Why only the 4 points? Should I not be trying to grab as much of the surfaces as possible to get the best representation at what is going on? I'm not arguing you, just trying to make sense of it all.

Please sign in to view this quote.

point spacing = Step width I assume?

I'm pretty sure I've done what you suggested and the attachments are of the results.

Please sign in to view this quote.

Ask and you shall receive, I am willing to give you guys whatever you need in order to help me figure out where my misunderstanding and issues lie. I am not a 'good enough' kind of person, I don't want to 'know' how to do something, I want to understand it. The monkey knew how to fly to the moon, but didn't understand it. I don't want to be the monkey. I know I have A LOT to learn, I just don't have the time or support where I currently am, no one to bounce ideas off of, which is why there's sometimes a long delay between my responses.

 

Thank you to everyone that is trying to help me.

Untitled21.png

Untitled22.png

Untitled23.png

Untitled24.png

Untitled25.png

Link to comment
Share on other sites

Possibly. Only tweak would be 0-380° - always overscan when possible as your worst points are your start and stop points. 

 

That Flatness plot for the most part looks amazing. The mag is set to 10,000:1, so everything looks extreme.

Link to comment
Share on other sites

Please sign in to view this quote.

Hmm, yeah I guess they maybe are. I have the start points at the normal +X/0°, and they go for 380°, so it looks like they go outward from the center of the part at about the 2:00 - 2:15 positions. I'll try to bump it up to like 450° and see if they move.

EDIT: Hmm, I guess I default the overscan to 380° on diameters, not the plane, my bad.

 

Untitled26.png

Edited
Link to comment
Share on other sites

Nominal and vectors in the Polyline can cause this effect, if they are not absolute zero the results will be based off the nominal value for Z. If your vectors in the polyline are not normal, it could have an effect.

 

 