Re-calibration of probe


Recommended Posts

  • 2 months later...
---

I, too, have been curious about this. I asked one of the application engineers during my advanced calypso training and they could not provide an answer at the time. 

In my experience with other CMM manufacturers and softwares, there is technically no true reference or correlation between styli unless they are qualified to the same instance of a master probe qualification. Like Richard Shoemaker stated above: there is little risk in just qualifying your single stylus and master again if there has not been a crash. The uncertainty in correlation between the styli will be likely less than a few microns. 

×
×