[Dy...] Posted 4 hours ago Share Posted 4 hours ago I could not find any clear answer on this, If I replace a probe and have to re find the ref. sphere location with the master to calibrate that probe do I have to calibrate all other probes on my rack or can I just calibrate the one probe I replaced and the rest will compensate? Link to comment Share on other sites More sharing options...
[Ri...] Posted 4 hours ago Share Posted 4 hours ago It depends.... Did you crash the machine? If yes, it's probably in your best interest to requalify all stylus/stylus systems. If you are just simply replacing a stylus due to wear or switching sizes/lengths, then most likely no. Link to comment Share on other sites More sharing options...
[Dy...] Posted 3 hours ago Author Share Posted 3 hours ago Please sign in to view this username. So nothing really saying you HAVE to do the others unless due to a machine crash. Link to comment Share on other sites More sharing options...
[Ri...] Posted 3 hours ago Share Posted 3 hours ago Not that I'm aware of. You should always perform some sort of routine qualification, but that frequency if completely up to you, and your company (usually risk assessment is needed). Link to comment Share on other sites More sharing options...