Please correct me if I’m wrong, and apologies if this has already been discussed.

When using the New GD&T, it appears the MMC bonus is being calculated based on the “suggestion” evaluation used in GD&T, which may not align with the evaluation selected for the diameter size.

For example, I generated a quick output where I left the diameter and true position at their default evaluations: LSQ for the diameter and “suggestion” for true position.

image.png.026c3988483b2b30e731c75443a23599.png
Based on the above, I calculated the MMC tolerance to be 0.2078. However, the report shows 0.1892.

Using CALYPSO’s calculation, the implied diameter would be 3.1592.

Listed Tolerance-Base Tolerance=Bonus

Bonus+Lower Spec=Actual Size

0.1892-0.03=0.1592

0.1592+3.0=3.1592

After reviewing the diameter evaluation and making an educated guess about what the “suggestion” evaluation was using internally, I applied an Outer Tangential evaluation. This resulted in the 3.1592 size that I was hoping for.

It seems the MMC bonus may be tied to the evaluation method used internally by the GD&T characteristic rather than the reported size evaluation itself.

I guess if you make perfect form holes/bosses, you've got nothing to worry about 😁

image.png.25a00d4ad17778372c54765ea9ab2516.png

image.png.bfb6bfb066309601144a60b2018a49b4.png

 

Link to comment
Share on other sites

Yes, definitely. If you are evaluating to an ASME print then Position should use Outer Tangential evaluation. Reported size should also use Outer Tangential, unless you are doing a capability study, then LSQ would be appropriate.

Link to comment
Share on other sites

image.png.0b306077c2ec69cbfdcce7b378b38d9f.png

 

image.png.47350581a40b9efada7f61a51258dc59.png

 

image.png.d02355dff4d5720ba25d1d191468d3ed.png

 

You can always change the evaluation parameters inside of the GD&T characteristic. In addition, always ensure you double check the tolerances for the MMC itself. If you copy/paste characteristics and change the feature via typing the MMC values will be incorrect. 