Jump to content

How are “Mean distance” and “Distance standard deviation” weighted in ZEISS INSPECT surface comparison?


---
 Share

Recommended Posts

I am evaluating internal gap using a surface comparison in ZEISS INSPECT/GOM.

For a virtually fitted implant component (local best fit with tolerance/outside target element), practically all valid deviations are positive because penetration is prevented by the alignment. Therefore, I expected the software-reported signed mean distance to be approximately equal to:

Integrated signed distance / Area of valid distance

and also approximately equal to:

Integrated absolute distance / Area of valid distance

However, in one example I obtained:

  • Mean distance: 0.0858 mm
  • Integrated distance: 31.1040 mm³
  • Integrated absolute distance: 31.1040 mm³
  • Area of valid distance: 175.5282 mm²

Thus:

31.1040 / 175.5282 = 0.1772 mm

which is about twice the reported mean distance, despite the absence of relevant negative deviations.

Could ZEISS please clarify the exact calculation of the following surface-comparison statistics?

  1. Is Mean distance calculated as an unweighted average of geometry inspection values, mesh vertices, CAD vertices, or another set of evaluation points?
  2. Are individual values weighted by the associated triangle or surface area?
  3. How is Distance standard deviation calculated, and does it use the same weighting as Mean distance?
  4. Is avg(inspection.value, index='geometry') mathematically identical to the Mean distance shown in the properties?
  5. Is there a built-in method to calculate a genuinely surface-area-weighted mean and standard deviation?
  6. Does the tessellation density or triangle-size distribution of the nominal CAD influence Mean distance and standard deviation?

My aim is to report the physical mean internal gap over the complete valid surface. At present, I calculate this as integrated distance divided by valid area, but I would like to understand precisely how this differs from the built-in statistics.

 
 
 
 
Link to comment
Share on other sites

Hi János,

the values are described in this article, together with the integrated distance: ZEISS Quality Tech Guide 

The Principles of the Surface Deviation might also be interesting.

I need to have a look at surface area weighted mean later, but it is possible to use the deviations and areas. Maybe a surface defect element might be used here since it has informations about area, deviation and points. With the python API it should also be possible of course. 

Nanno

 

 

 

 

Link to comment
Share on other sites

Thank you for the quick response. I knew those guides, but I want to double-check my understanding. I have already made a custom add-on for MAD = Integrated signed distance / Area of valid distance.

Here is a screenshot of our project. The discrepancy probably comes from the density of the triangles. Areas where the surface deviation is minimal are very dense, so they are weighted more in the mean distance (AVG) parameter because there are more points within the same-sized area. Whereas the integrated distance calculation weighted all triangles equally, as it was controlled by their areas. Am I right?

image.png.4a12309982cf70143f902f744df6dd84.png
image.thumb.png.a94c4dbab16fdaa4fd694ccc59340a87.pngimage.thumb.png.cddd043e5e1eae8a378cde83dfdacaab.png

Link to comment
Share on other sites

Yes, the large triangles will represent the surface deviation with lower resolution, while the surface area stays more or less the same. 

In the properties of the CAD or during import you can define the surface tolerance and max. edge length ZEISS Quality Tech Guide

Alternatively, you might use a surface comparison on actual (on the mesh with higher resolution). The CAD should be refined anyways. Keep in mind that the integrated distance is an approximation of the volume, so maybe you want to analyse this with additional inspections. Maybe continuous curve distances using sections on CAD and mesh? ZEISS Quality Tech Guide

Link to comment
Share on other sites

 Share

×
×
  • Create New...