[Pa...] Posted August 11 Share Posted August 11 Hi all, hoping someone's run into this. I keep getting a warning "Caution, stylus not qualified no. 5" The catch is my active stylus system only has one stylus, stylus 1. So there's no stylus 5 to qualify. The warning only pops up after I remeasure a feature, not on load or qualification. That makes me think a few features are carrying an old "stylus 5" assignment in their strategy how can i get rid of this? Link to comment Share on other sites More sharing options...
[Ri...] Posted August 12 Share Posted August 12 When the message comes up, in the top right of Calypso should be the feature name. You should experience a Red light condition, DO NOT ENABLE THE GREEN LIGHT. Doing so will remove the feature name at the top right as I mentioned. Go to that feature and remove the offending stylus number. 2 Link to comment Share on other sites More sharing options...
[Gr...] Posted Thursday at 02:28 PM Share Posted Thursday at 02:28 PM Sounds as if this is the same issue I was having when running 'Single Char.' and 'Executing'. Turning off "End Park Position" works. Something to do with ending with a different probe than the saved park pos. 3 Link to comment Share on other sites More sharing options...
[Br...] Posted 6 hours ago Share Posted 6 hours ago Greg is on the right track, A park position is sometimes made with a stylus or stylus system that no longer exists. So, if you look in the Measurement Plan Editor>Stylus, you will not see the error. Removing this stylus in there and choosing an existing one will correct the issue. Link to comment Share on other sites More sharing options...