We have programs made for running qualification checks on the probes that output the sigma value to a report.  Once in a while a probe branch will not pass qualification and still output a "passing" sigma.  The sigma for each branch is output with a Result Element formula getProbe("Down","3 MM X 50 MM").stdProbeDev.

Obviously probes not qualifying is an issue in itself.  Are there ways to check the pass/fail status within a program like this?

  • Like! 1
Link to comment
https://qualityforum.zeiss.com/topic/27429-reporting-qualification-results/
Share on other sites

Please sign in to view this quote.

Hello Daniel, 

 

you are right, the current situation within CALYPSO is, that as soon as the "Sigma" Value Limit is reached, the "old/good" Sigma value from the previous qualification is used in the Result Element. 

The Reason why the faulty qualification doesn't overwrite the good qualification is that it could cause a vector change that could lead to a collision in the following qualification.

Development is working on a solution for a better visualization.

Currently you can only see that the Stylus is not qualified in the Stylus System Management (marked as red) as well as the "default report". The default Report shows that the stylus is not qualified and additionally the correct/bad Sigma value.

 

Best Regards