[Ro...] Posted 5 hours ago Share Posted 5 hours ago Hello, Currently based on the number of stylus systems we have defined we must take a two-step approach to qualifying all of our probes on each of our Contura's. Step 1: traditional Z+ orientation, qualifies 95% of the probes in our probe system library (rack) Step 2: angle the calibration sphere 90° and repeat for the Z+ probe we have The Question: Is there a method of defining two calibration spheres in a single instance of a Calypso qualifications program in order to learn/place two calibration spheres on the table at one time and run the probe qualification in a single program run cycle without the need to refixture the qualification sphere? Please paste a link if this is already explained somewhere else, could not find anything based on my search words Link to comment Share on other sites More sharing options...
[Ma...] Posted 5 hours ago Share Posted 5 hours ago We have in 45° in both directions, so for calibrating all 5 probe tips we are running also in two stages. Perhaps using another angles would reduce your two positions. Link to comment Share on other sites More sharing options...
[Ro...] Posted 4 hours ago Author Share Posted 4 hours ago Please sign in to view this quote. I would think Zeiss has many requests similar to this and would have a fixture featuring multiple directions just for this scenario but so far no luck... Link to comment Share on other sites More sharing options...
[Ma...] Posted 3 hours ago Share Posted 3 hours ago For answer - everytime you move your cal. sphere you have to find it's location via REF. probe. We have original stand like this one. We just rotate by 180° in Z axis to cover rest of probe tips. What configuration do you have? Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in