Does anyone here know if the fix assist has the ability to turn a probe to 120 degrees? I have a 6 way hub with the styli threads at 60 deg apart. I need to build out a system where the styli are at 120 using said hub I bought. It will end up being a 3 way system. I cant remember if the fixassist will do that or not.

Also, what master sphere would you guys recommend for calibrating a 0.5 mm probe?

 

Mike

Link to comment
https://qualityforum.zeiss.com/topic/27706-fix-assist/
Share on other sites

Hello,
We use an 8 mm sphere to calibrate our 0.5 mm or 0.3 mm probes. Whether a larger sphere can be used depends on the length of the shank. The problem is that it is difficult to tell whether the shank has collided with anything during calibration: the parts are too small to be visible.
That was the reason we decided to use an 8 mm sphere.

Link to comment