Prismo MPE


Recommended Posts

In real life, if everything is working properly, then you'll always have a much smaller length measurement error than the stated MPE value.  If you look at your calibration certs, you'll see they are well under the stated MPE values.  I take care of my machines and that's the case for me at least.  

 

That's why the Prismo Ultra will advertise it's capable to 0.5 + L/500 µm, but its stated MPE is only 0.9 + L/350 µm.  

Link to comment
https://qualityforum.zeiss.com/topic/27728-prismo-mpe/#findComment-157358
Share on other sites

I am not sure if this Prismo is the Ultra or they are separate systems.

I have worked with the large Prismo Ultra's at Cummins in Rocky Mount, and I swear those calibrations were under 1.0 for 1800 sizes. I could be wrong.

Hell, I may be reading this report wrong, I usually don't pay them much mind.

As far as I can tell, this particular CMM came in around 1.5, I thought it would have been lower.

Link to comment
https://qualityforum.zeiss.com/topic/27728-prismo-mpe/#findComment-157370
Share on other sites

×
×