[Ke...] Posted 1 hour ago Share Posted 1 hour ago Does the "Curve Distance" evaluation(s) use actual points or the calculated feature for results (for instance, if 4000 points was used on one curve and 5000 points was used on the other)? Also, what if the Curves were scanned in opposite directions, or with different start/stop points (in a 360 degree scan) Link to comment https://qualityforum.zeiss.com/topic/27758-curve-distance-evaluation/ Share on other sites More sharing options...
[Ri...] Posted 1 hour ago Share Posted 1 hour ago It "should" be the curve corrected points. I've had issues when the two curves were setup differently, so typically I just define one of them, copy, and offset the 2nd one, so that they match. Usually the start/stop isn't much of an issue, it's typically the point numbers being drastically different, so there isn't a directly opposing point for calculation. 1 Link to comment https://qualityforum.zeiss.com/topic/27758-curve-distance-evaluation/#findComment-157486 Share on other sites More sharing options...