[Ja...] Posted October 12, 2023 Share Posted October 12, 2023 In the attached image I've used a symmetric plane to report out size. I understand the standard in reporting out the Max and Min to consider the entire envelope. However i'm only interested in the max size in this case. Is there a way to only report out the max? Link to comment https://qualityforum.zeiss.com/topic/3320-linear-size-is-there-a-way-to-separate-minmax-size-or-only-report-out-one-measurement/ Share on other sites More sharing options...
[Mi...] Posted October 13, 2023 Share Posted October 13, 2023 You can open the linear size dialog and chose "(GN): Minimum circumscribed element" for both the upper and lower limits. Link to comment https://qualityforum.zeiss.com/topic/3320-linear-size-is-there-a-way-to-separate-minmax-size-or-only-report-out-one-measurement/#findComment-9092 Share on other sites More sharing options...
[No...] Posted October 26, 2023 Share Posted October 26, 2023 (edited) To seperate min and max, select the linear size element, then go to Inspection -> Check Dimensions -> Other -> User-Defined.. Under "Actual Expression" type inspection followed by a dot: inspection. When you type the dot, a context menu should open. In the menu select Results -> result_dimension_min_max -> min -> measured value This will give you the minimum value in a separate element. For the max value repeat the above, but select max. Edited October 26, 2023 Link to comment https://qualityforum.zeiss.com/topic/3320-linear-size-is-there-a-way-to-separate-minmax-size-or-only-report-out-one-measurement/#findComment-9241 Share on other sites More sharing options...