I am using the GomScan1 200 lens device. When I scan, I get the error you see on the screen. What is the reason? What is the solution for this problem?

 

image.thumb.png.7a5ab2855553fc258eee48ecb3c88023.png

Link to comment
Share on other sites

No. I use store curtains in my office. But I don't work as a single person in the office. Does movement affect ?

Link to comment
Share on other sites

I guess those let light through? The message says that the lighting conditions change during the measurement. Movement is not an issue here. Can you test it in a dark room to check if the light that goes through causes this issue?

Link to comment
Share on other sites

Hey. try to change your Kamera setttings.

Kamera > Qualität > Userdefined   . Set a spot and press the button.

You can set the Focus ,so the software optimizes the light for the spot you set.

Maybe someone changed /played with the settings of the Kamera.

kamera.JPG

kamera2.png

Link to comment
Share on other sites

Hi,

from my point of view these kind of questions couldn't be solved properly via the Forum. I would strongly recommend that you contact your local ZEISS partner for a direct support!

Regards

Link to comment
Share on other sites

  • 2 weeks later...

Hi

 

I´m having the same problem, I will tried to fix it with the sensor calibration and if this solve, I will let you know

 