Measurement error


Recommended Posts

I was trying to measure a ring gage to correlate with our ULM system when I noticed that the dimensions were way different.

Ring gage should have measured 2.99996 and CMM is measuring it at 3.00046 a difference of .0005.   Have also checked gage blocks and getting error within .0005 - .00076.

Machine is a Zeiss Prismo with a Vast Gold head.  I have verified that the correct head is selected, the calibration sphere dimensions are accurate, requalified Master, and probe used for inspection has been calibrated as well.

What am I missing?   Is there something else that could be causing this error?

Link to comment
https://qualityforum.zeiss.com/topic/4400-measurement-error/
Share on other sites

  • 2 weeks later...

Please sign in to view this quote.

You are using the R value from the Zeiss certificate right? That's quite the error at 0.0005". Please list measurement strategy and evaluation settings, although with an error that high I would think there is physical damage to your equipment. Whats the probe sigma like?

Link to comment
https://qualityforum.zeiss.com/topic/4400-measurement-error/#findComment-121248
Share on other sites

×
×