[St...] Posted October 10, 2023 Share Posted October 10, 2023 How do you know when it is time to replace a worn out probe? Have an operator that says he can see the flat spots with his naked eye and wants to change probes. So how long do Ruby and Silicon Nitride probes last? The sigma on the qualification is less than .0005 mm.Probe Qualification.JPG Link to comment Share on other sites More sharing options...
[Da...] Posted October 10, 2023 Share Posted October 10, 2023 Ruby is an extremely hard material, only slightly less hard than diamond. In my opinion, you will never see a flat spot on a ruby. I've checked a lot of probe rubies under a 2000x Microscope and never have I seen a flat spot. In fact, I had been searching for a perfect way to clean aluminium build-up on ruby, and I've saved and improved a lot of ruby probes by using sandpaper. Yes, you've read right, sandpaper. The ruby doesn't even shed a smile when I torture it with sandpaper, and it will get the range (not the sigma, but the range, the sigma will even be better) back to values under a micron. One thing that I did see on rubies is pitting. When I have a bad ruby, it's always pitting. I've had a probe ruby used on a Heidenhain probing system on a CNC mill that was totally unusable because it looked like huge chunks were broken out. My suspicion was the operator had dropped it on a hard surface, no other way it could have happened. Virtually all rubies I've checked have small pits that are small enough to not affect everyday measurement. Even brand new rubies have those. Every small pit is a potential nucleus for a material break-out, but if you treat your probes right, they can last for years. My worst calibration results happened on new probes and could be explained by faulty glueing, thus resulting in almost unnoticable movement at the connection to the stem. My small vendor replaces those without lifting their eyebrow. If you do heavy work on abrasive materials all around the year, even then will you most likely only enlarge the pits that had been already there but not create flat spots. Try it, take an old probe and try to destroy it with sandpaper, and check before and after. Yes, there may be tiny scratches visible under a microscope, but their width is negligible. If you are unsure, just measure your reference sphere with a scan around the equator and some scan lines on the rest of the surface. The most stressed part of a ruby probe usually is the equator. Any defect will reflect right away in the scan. As for silicon nitride, I use some probes, but I have too little first-hand information about wear. I will report - in a few years? 😃 Link to comment Share on other sites More sharing options...
[Ri...] Posted October 10, 2023 Share Posted October 10, 2023 Please sign in to view this quote. I have been using electric erasers for about 20 years, I use Mylar eraser inserts, they do no harm to ruby or Silicon Nitride. Typically, I can clean rubies from 2mm on up and disc styli using a 10x scope or eye loupe. This can clean off aluminum, magnesium, titanium, etc. build up within about 10 minutes per stylus. Link to comment Share on other sites More sharing options...
[Ke...] Posted October 10, 2023 Share Posted October 10, 2023 It depends on what material and what the condition of the surface you are measuring as to how long probes last. It is definitely possible to flat a Ruby, Silicon nitride, or even Diamond-coated probe (sort of. Usually the diamond coating breaks away and the underlying material gets flatted in that case). It is also possible to have buildup occur on the probe in such a way that it appears to be a flat. I personally have customers that have run the same ruby probes for years, usually not having to replace them until an operator incident breaks them, and I have other customers who have to replace the probe every 1.5-2 months due to flatting because all of their measurement is done with the end of the probe, on surfaces with machining marks, at very high measurement volume. One way to catch these is to create a program which scans the reference sphere as a circle, or semi-circle depending on access, in multiple axes, then display and plot roundness with some level of magnification (I believe I used 500x for the customer in question). This will allow you to visually identify using the plot if there is either build-up or flatting. Link to comment Share on other sites More sharing options...
[Ri...] Posted October 10, 2023 Share Posted October 10, 2023 Did I just read that you will never see a flat spot on a ruby??? Lol. Someone hasn't scanned a lot of planes on semi-rough surfaces before then. Link to comment Share on other sites More sharing options...
[Da...] Posted October 10, 2023 Share Posted October 10, 2023 Please sign in to view this quote. Pictures or it didn’t happen! 😉 Link to comment Share on other sites More sharing options...
[Ri...] Posted October 10, 2023 Share Posted October 10, 2023 Please sign in to view this quote. Lol. I don't have them anymore, but it absolutely happens. Scanning planes is the worst wear on probes due to the localized surface contact. Link to comment Share on other sites More sharing options...
[Je...] Posted October 10, 2023 Share Posted October 10, 2023 Please sign in to view this quote. Probes are relatively cheap. When I encounter any plausible existence of wear or failure, I replace the stylus. The time and risk in reworking a worn stylus is greater than the time and cost it takes to replace one. If an employee approaches me with doubt over a stylus tip's quality, I immediately replace the tip with no questions asked. I thank the operator for being proactive and communicating about the problem. Stat. models and PM could be used to detect stylus failures, but these would only add value in large volume situations. Link to comment Share on other sites More sharing options...
[Da...] Posted October 10, 2023 Share Posted October 10, 2023 Flat spot on a 5mm ruby caused by scanning ceramic. I have an image from a stereo microscope somewhere also.. it can happen. 😃 Yellow line is ‘perfect’ geometry and you can see the gap to the tip.IMG_3328.jpeg Link to comment Share on other sites More sharing options...
[Ow...] Posted October 11, 2023 Share Posted October 11, 2023 Been working with CMM's for 18 years and yes, you most certainly can use a probe enough to wear a flat spot on them (especially on cast iron parts working three shifts 5 days a week) and a standard tensor or just points probe calibration will not detect them, especially on the very bottom. Silicon Nitride probes are best for aluminum, not so much for wear in my opinion but, because you can see the aluminum build up on them. Link to comment Share on other sites More sharing options...
[Ro...] Posted October 11, 2023 Share Posted October 11, 2023 If everything has been calibrated together, take a Z point on a very small surface of something, then change to a different probe and take a point in the same place. if both points are within a couple tenths of each other, then tell your guy to get back to work. (do things like this while your operator is at lunch. if the measurement is off by a substantial amount, or off in the wrong direction, it'll just lead to more questions and stress in your life) Link to comment Share on other sites More sharing options...
[St...] Posted October 12, 2023 Author Share Posted October 12, 2023 I would like to thank everyone for their input on this subject. It is very much appreciated!! We do quite a bit of scanning on steel as well as aluminum parts. So I can understand that a flat spots can happen, and that the probes do wear out. But as I say we only check the "shiny" surfaces. So the surface is not too rough to really wear the probes out real fast. Link to comment Share on other sites More sharing options...
[Ch...] Posted October 12, 2023 Share Posted October 12, 2023 Use this program to find flat spots on rubies. This was originally from Derek Stedman.Styli Roundness Scan.zip Link to comment Share on other sites More sharing options...
[Ro...] Posted October 13, 2023 Share Posted October 13, 2023 This is after scanning a part made of silicon carbide. It ate up the Ruby, so then we switched to Silicon Nitride which wasn't as bad as the Ruby but still wore it down. We are going to try using Zirconia but if that fails then Diamond it is. Link to comment Share on other sites More sharing options...
[Gr...] Posted October 17, 2023 Share Posted October 17, 2023 I also work with silicon carbide and see a similar thing on our ruby tipped probes. We usually swap out a probe when the stand deviation is over 1 micron, and it's almost always a flat spot on the bottom. Our 0.5mm probes last about a week before there's a flat spot on them. but the 3mm probes can go on for months or more. And yeah, we could switch to diamond coated probes, but they're literally 10x the cost of a ruby tipped probe. Link to comment Share on other sites More sharing options...
[Da...] Posted September 16, 2024 Share Posted September 16, 2024 I agree with Owen, I've been programming CMM's for 46 years and yes i started before Computer Monitors Mice Hard Drive or RAM, we had a Teletype Machine with Paper Tape Punch, you can use a probe enough to wear a flat spot on them. when i went to another company after they lost a head programmer i ended up scrapping 6 of the stylus tips some you could see Flat spot on tip & Around the equator where they were scanning brass rings I.D. & O.D. - - - the machine shop complained the CMM's didn't render same results as Deltronic Pin Gauges i knew right off i needed to check the stylus tips and make sure the programs were using the proper evaluation settings . . . . 😮 Link to comment Share on other sites More sharing options...
[Wo...] Posted September 16, 2024 Share Posted September 16, 2024 Please sign in to view this quote. This is my approach as well - if in doubt - replace it. The risk is just not worth it. I also managed to flat-spot a probe. Curiously, the sigma was perfect as the Tensor calibration is not hitting the top of the sphere, and that's where a flat spot was (about 0.002" deep) Link to comment Share on other sites More sharing options...
[Mi...] Posted September 17, 2024 Share Posted September 17, 2024 For regular steel parts our probes often last years and that's in a high volume environment. Generally they get replaced because they are broken, or glue has failed. We do have a a nasty scanning job on a rough ceramic disc that eats up probes fast, like above we considered diamond probes, but the decision was made to get the shortest, cheapest ruby probe and just replace them every few weeks. The job only measures flatness, which stays accurate enough for it's purpose even with a noticeable flat spot. Link to comment Share on other sites More sharing options...
[Lu...] Posted March 24, 2025 Share Posted March 24, 2025 Hello, what are your all's thoughts on this ruby? I saw how this ruby has 2 rings on it and i wanted to know if this would warrant replacing it. Im new to the cmm world so i am unaware if these rings would indicate any loss in accuracy or if they are pretty common. For reference the ruby is 5mm in diameter. Thank you all very much!  Link to comment Share on other sites More sharing options...
[Ja...] Posted March 24, 2025 Share Posted March 24, 2025 (edited) Please sign in to view this quote. Luke, Looking at that ruby I would err on the side of replacing it but you can try gently wiping it first with a clean microcloth and some isopropyl alcohol to see if those "rings" clean off first. My guess is those "rings" are forming from scan paths you are consistently using and beginning to wear into the ruby so they won't simply wipe away. The probe may still be qualifying fine and you may not be seeing loss of accuracy yet however like others in this thread have stated it is usually safest and easiest to swap it out for a new probe when there is any doubt. Edited March 24, 2025 Link to comment Share on other sites More sharing options...
[Ri...] Posted March 25, 2025 Share Posted March 25, 2025 Please sign in to view this quote. That would clean off quite easily using the method I posted earlier about electric pencil eraser and Mylar inserts. Link to comment Share on other sites More sharing options...
[Ma...] Posted March 25, 2025 Share Posted March 25, 2025 It looks like you have been scanning aluminum. I think it is material built up on the stylus rather than wear. If this is the case, it may be prevented by using a different material than aluminum-oxide (ruby) such as silicon nitride for scanning aluminum. Link to comment Share on other sites More sharing options...
[Lu...] Posted March 25, 2025 Share Posted March 25, 2025 Thank you all very much i appreciate the responses Link to comment Share on other sites More sharing options...
[Da...] Posted March 26, 2025 Share Posted March 26, 2025 As far as when to replace the styli, I'd say it depends on the tolerances of the parts you work with. If the tightest tolerance on a workpiece is +/- 0.25 mm then throwing out a stylus with a 2 micron qualification sigma would not be advisable, however if your tightest tolerance was +/- 0.01 mm then you may want to replace it. Link to comment Share on other sites More sharing options...
[Ma...] Posted March 28, 2025 Share Posted March 28, 2025 Please sign in to view this quote. Erasers for erasing on mylar surfaces? or like a full blown mylar eraser. Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in