Z+ stylus qualifcation
We are adding a four sided star probe to our arsenal and have ran into a snag when qualifying the Z+ position.
Our reference sphere points up and out 45°. Similar to the one below.
https://shop.metrology.zeiss.com/INTERS ... ation=true
This won't work for qualifying Z+ since it runs into the shank. My thought is I have to go to something like this
Does anyone have an alternative?
Thanks in advance!
Standard RSH Sphere adjusted down to 120° instead of the default 135° should let you calibrate +Z Probes we do it everywhere.. XXT and XT.
  • 2 weeks later...
There is also a 6mm bolt on the back of the RSH that will allow you to rotate its "head" which will allow you lay the Reference Sphere over to 90° of tilt. This will give you the option to qualify all of those tips with one Reference Sphere position. You will need to rotate the base though to make it parallel with the XXT.
Yes but only one of our machines actually has the fully built left and right sides of that system and it's calibrated constantly. Most of the production machines only have the left or right set of probes, and even then they often only half a couple of them. In simulation I program with complete stylus systems so I can move programs around to different machines and never have surprise crashes because I didn't account for a probe that isn't there on the original machine, but is present on another the program may get moved to.
90°/0° is absolutely better for that +Z Probe. The reason I said 120* and not all the way down to 90* is the RSH sphere is harder than it needs to be to adjust. 120* will let you calibrate all your regular probes on a standard star, and a +Z probe. So the tilt never has to change, just rotate the sphere, remaster and continue calibrating.
No.
You were originally at 135°, 120° is only down a bit.
That's an RSH Sphere, do you actually use the RSH function? It makes adjusting the sphere position so much easier. One of my favourite discoveries when I was new.
What we have done for the last 12 years is have the reference sphere at 135°, put it 45°X+Y+ on the CMM to get the majority of our qualification program and swivel 90° to get the rest. So we do use the RSH function but in a very limited scope.
This four sided star is the first time we haven't been able to get everything needed with the method described above.
We have a second reference sphere in the lab, I am working on combining them into some combination so that we can get every position needed, see attached.
You're actually not using the RSH, it's greyed out "RSH Shaft Definition". I suggest you set it up as it'll make this job you're doing easier. It's a built in way to make the CMM measure the Sphere tilt and rotation.
Make a new sphere with the correct RSH components 3535_50fb22d9e6991a6202aa86c4dd69fc58.png
Then you can do this and get your angles exact. 3535_c673ef2be6461d953ddedbeb1ca60026.jpg
But yes, 120°45° (or 135,225,315) and 120°/0°( or 180) should allow you to calibrate a very wide range of probes without ever having to adjust the tilt which kinda sucks to do on the RSH if it's a regular job. 3535_f69de756ea08342ee177df0c346cf68a.png
  • 4 weeks later...