When using characteristic conditions,The result is masked. I need help
Hello, everyone !
I have a question.
1. I use a conditional function on the characteristic, which is set to remeasure if out of tolerance.
2. However, when the result is in tolerance after re-measurement, the result is not output as a PDF for only the relevant characteristics.
(Other characteristics are output as PDF.)
Tolerance was entered in the characteristic conditions,The setting repeat_until is used.
Can I include the re-measurement results in the report?
Hello.
If the measurement falls within the characteristic tolerance, the result is output.
However, if you remeasure with out of tolerance, there will be no output.
I wonder if there is a way to reload the current results after the measurement is finished and have them automatically saved
If I copy the characteristics and set the opposite conditions, is it possible to print it?
I'll try it tomorrow.
Do it all on the Feature side and this problem will likely vanish, also will save time as it will remeasure the feature immediately instead of at the end of the run.
Get rid of the condition on the Characteristic.
The example below is all on the feature, it will immediately remeasure the hole up to 3 times if the size it +/- 0.013mm from nominal. I actually find it to be more useful to use a form limit than a size limit on newer programs.
No position moves, or operator input required. They don't even know it happened unless they stand there and watch it run.