Ruby Diameter vs. Sigma Value


Recommended Posts

I only say toward "larger" diameter probes for certain tasks.
Smaller for form measurements, as Clarke mentioned.
My most common used diameters are 1mm & 3mm.
Lower sigma values are definitely preferred.. It would depend on how big of a difference are we talking about, and the tolerances are for the measurement task.

Please sign in to view this quote.

.
Both stylus systems are intentioned to give a plausible, accurate result. If either has an inordinately large sigma value, visually inspect the probe under magnification and make sure it is fastened securely. Are you measuring size such as a diameter or form such as flatness?

In situations where the measurement task is to evaluate compatibility of a mating part, such as an inner diameter, I often use a larger probe and evaluate using outer tangential element. If I need to evaluate form deviation such as flatness, then I often use a smaller probe.

You are probably wanting a simple answer such as "larger probes are more accurate" or "smaller probes are more accurate," but neither answer is true. The correct answer is "use the right tool for the job," or in other words, "use the right sized probe for the measurement task."

This being said, the 1.35mm probe has a thinner shaft, and I have dealt with deflection issues in the past when pairing a 1.35mm probe in combination with longer shafts. If you have a VAST sensor, decrease the measurement force and dynamic probing. Regardless of which sensor you have, decreasing the scanning speed may help.
.
  • 9 months later...

Please sign in to view this quote.

hi,
im having problem with my probes (VAST XXT) and i think the force im using is too much, have tried to lower dynamic % but it didnt do anything 100%, 50% and 10% give me same results

how can i change the measuring force?

i know this is an old thread but hope u still on 😃

Please sign in to view this quote.

The XXT cannot generally control the measuring force electronically. You can switch to Sensitive, or Robust to slightly decrease/increase the force, but the change is minimal. If you truly need to make the force lower, you have to do it mechanically, by making the stylus system longer.

The % values that you are speaking of are the Probing Dynamics they do not control measuring force. They control the percentage of speed that the machine approaches the surface once inside the retract distance.
659_24a648a2542870e65e1ec512e986c386.png

Please sign in to view this quote.

ok maybe i understood measurment force wrong,

i have a problem with VAST XXT, after updating from calypso 21 to 22
which was a bad mistake for the amount of problems it has compared to 21

my biggest issue now is when i try to calibrate my probes (7probes) i got 4x the Sigma value i was getting before,
the only S value that didnt change is the masterprobe
and i thought that Calypso 22 changed some settings or so

any idea why i get high S values?

i have tried almost everything but nth helped
the probes are clean and not damaged at all

Please sign in to view this quote.

so this one had high sigma value even before
its an L probe with long shaft

normally i would get 0.0185

but now i get around 0.08 i measured this one specifically 4 or 5 times, same results

other 6 probes i was getting 0.002 now it’s around 0.005 - 0.008

[img] https://postimg.cc/xJx3PFQv[/img][/img]
AHHHHHHHHHHHHHH!! I see now. This is CAA Qualification. Yeah - sigma value standards kind of go out the window with a CAA Qualification.

The quality of the qualification purely depends on length, but even then there isn't a published formula to calculate what it should be.

I will say that 80 microns is a lot, but I've heard upwards of 100 is the limit.

I think your L probe is too long as well (axially).

To test and see if it a CAA issue, you can create a new Stylus System with that same buildup that isn't CAA enabled and see what the sigma values show with a standard qualification.
×
×