What do you do if .arn was overwritten by a part program? Auto Run is not keeping the Layers we are having to open the program folder every time. I suspect the .arn was overwritten could be causing this. The only one I see is in inspection folder. Hopefully there is another place, and this is not the issue. If it is not, I am at a loss for why the Layer drop down is not working. This is on a brand new Contura.
Link to comment
https://qualityforum.zeiss.com/topic/7788-74-auto-run/
Share on other sites

  • 1 month later...
  • 3 years later...

I noticed that on one of the machines I open auto run from the Start desk and it automatically goes to the .arn file and on the other machine it opens autorun and I have to tell it to open the specific file for the autorun to open, does anyone have any experience with that? So, one machine loads the .arn file automatically and the others go to this and I have to file, open and tell it the .arn file. Not sure why though.

Full screen snip for Autorun.PNG

Link to comment
https://qualityforum.zeiss.com/topic/7788-74-auto-run/#findComment-155250
Share on other sites

To load an autorun file automatically, first set Calypso to start with autorun. While doing so you can select an *.arn file to open in the path field below. After selecting the file, switch the setting back to normal Calypso. Now whenever you open autorun, it should automatically load the *.arn you put in the path field.

But beware, this setting has to be done seperately for each Calypso user, and you should be prepared to lose it again after certain update procedures.