Fourier analyses
I haven't done this myself, but it should be pretty straightforward using the waviness evaluation. You put the values into the dialog, check "double amplitude", and you need a certain amount of points (300 waves per rotation max for the Fourier analyis means: 300 upr * 7 points per wave = 2100 points for each evaluated circle minimum, if you use a cylinder you need 2100 points minimum per section). Use the smallest probe diameter possible and slow speeds.
The shown formula "RTA curve" is what you need. I think you need to limit the values in Harmonic and I'm not sure about the Type of Evaluation, but maybe you can take it from here.
Thank you very much 😉
About number of points, the main thing is to have at least 2100 points per circle, it is good like in attached picture?
Only thing I don`t understand is about "Upper and Lower Limits"? What should be there? How you understand it?
What does this evaluation shows to me in overall?
Captu1re.JPGFetching info...
The number of points is good.
The drawing calls for all harmonics <= 300 waves per rotation, so 0 and 300 should be your limits. That’ll give you a pass or fail for every harmonic, not only 2 as you have now.
See where the waviness goes down and then up again from around 270 upr? That’s the range where your waviness might fail.