[Ri...] Posted April 24, 2023 Share Posted April 24, 2023 Hi all, It appears that if I import an SSC and turn on the CAA function the offsets are getting compounded onto the stylus system. This makes it impossible for me to test/program with an CAA enabled stylus system. Link to comment Share on other sites More sharing options...
[To...] Posted April 27, 2023 Share Posted April 27, 2023 I can remember many years ago, that a CAA stylus system would not behave properly until you ran it through a qualification sequence. Haven't seen that in a long time though. Link to comment Share on other sites More sharing options...
[Ri...] Posted April 27, 2023 Author Share Posted April 27, 2023 Please sign in to view this quote. Tom, I love you. I ended up having to Qualify the MasterProbe, and perform a Fitting Position. Then on the CAA Stylus System, I set the XYZ to the correct values, and then performed a qualification. It now works correctly. Where were you a few days ago?? Many thanks as always. Link to comment Share on other sites More sharing options...
[Ri...] Posted April 27, 2023 Author Share Posted April 27, 2023 I just tested again. It appears that I only need to Qualify the MasterProbe, and perform the Fitting Position, and newly imported CAA Stylus Systems behave correctly (no need to correct the XYZ nominals). Link to comment Share on other sites More sharing options...