Hello,
I've used the Probing System Qualification to create a program to calibrate all of our probes.
I'm outputting the results to a pdf using the Default Report format.
By default, the results are going to C:\Users\Public\Documents\Zeiss\CALYPSO\workarea\results.
I want the results to go to a different location so I created the path shown in the screenshot.
This is working except for the file name is ".pdf" and the results overwrite each time.
I would like to have the program name (Probe Calibration) and a new file created each time the program runs.
Any ideas?
Is there an easier way?
2707_bf7a612f780705c8cd314218859fefda.jpg
Link to comment
Share on other sites

Put the date and time in the File name.

Right now you're creating a file with no name.

getRecordHead("planid")+"_"+getRecordHead("date")+"_"+getRecordHead("time")+"_.pdf"

or

"static_file_name_"+getRecordHead("date")+"_"+getRecordHead("time")+"_.pdf"
Link to comment
Share on other sites

I would recommend setting up a Name for all measurement plans using variables. This allows you to tick the box in any program and will create a file folder and file name according to your needs. You will never need to map anything else again.

name.JPG

Link to comment
Share on other sites

Please sign in to view this quote.

This is actually a good idea; I'll be implementing this with our programs. Though with multiple machines, may want to add a serial number in there, or some sort of CMM identifier.