[To...] Posted November 9, 2022 Share Posted November 9, 2022 Can someone possibly explain how to do a Go/NoGo or Pass/Fail manual entry in PiWeb Designer? Link to comment https://qualityforum.zeiss.com/topic/9226-gono-manual-entry/ Share on other sites More sharing options...
[b2...] Posted November 9, 2022 Share Posted November 9, 2022 You'll create the characteristic in Planner first. Create the characteristic and then: Change the characteristic attribute "Characteristic Type (K12004) to Attributive. Change the Measured Value Catalog to Measurement Status Catalog. The Measured value catalog is the responses that the operator or inspector will see as their options (Pass / Fail, or Accept / Reject, you can change the verbiage to whatever you want). Then in your report, you can simply use the Text Input Field connected to Measured Value and it will flip between expecting a typed number or a dropdown of Pass/Fail for attributive characteristics. That's the easiest response, there are obviously more ways of creating this in designer. Link to comment https://qualityforum.zeiss.com/topic/9226-gono-manual-entry/#findComment-42386 Share on other sites More sharing options...
[To...] Posted November 9, 2022 Author Share Posted November 9, 2022 Thank you for the assistance, Laura. Link to comment https://qualityforum.zeiss.com/topic/9226-gono-manual-entry/#findComment-42388 Share on other sites More sharing options...