Hello,

Recently we had a slight problem while we ran a program that I made in offline mode.
When I used star-system that was definied in offline too and I wanted to run it on the CMM it almost collided, the machine "thought" that the 0_90 (+X) stylus is located in the middle of axis Z.
As I said It is definied in offline as X56.75 Y-0.003 Z-7.903 R0.75 Xvector 1, the others are 0

Could you tell me what I possibly did wrong?

Thank you 🙂
Link to comment
Share on other sites

Did you create the master probe correctly in simulation? It's length must match the physical CMM.
Which type of probe head do you have?
With a VAST/XT head my preferred workflow is as follows:
* Select and save the correct master probe in Stylus Creator (needed only once)
* Import master probe in Calypso Planner (only once)
* Construct stylus in Stylus Creator and save
* Import stylus SSC in Planner
* Make and test program
* Export stylus from stylus management window (file format PCF)
* Import stylus PCF at CMM
* Calibrate with caliper button (works only on VAST / XT heads) or manually.