Calibration process - with error message


Recommended Posts

Hello GOM team!

 

I was wondering if you can help me with my problem. 

 

Failure is pictured below.

image.png.9e3fc276553dd8dd633a48246481e53c.png

 

This happen everytime i am running a calibration on our ATOSscanbox 4105.

I am able to run the calibration up to the C2 position where i get this error.

image.png.26a03d98622206fff171888a601654d9.png

 

is this connected with the exposition time setting or the distance from scanner and Calibration panel?

 

Thanx for the troubleshooting!

 

Dusan

 

 

Edited
missing text
Link to comment
https://qualityforum.zeiss.com/topic/993-calibration-process-with-error-message/
Share on other sites

Hello. If it is ok in one position and not in another with this cannot find elipses error my first thing to check would be dust or mark on lenses.

If the dust is covering identification elipses the error will happen.

Look carefully at the camera images in the software on first position and error position.

Depending on sensor variant start with cleaning lens cover/filters first.  If still not ok clean lenses . If still not ok ccd.

  Always use the manual air puffer.

Hope this helps