Hello, Recently we got a new computer for our Contura. Since then we have been getting inconsistent measurements. After qualifying the probe we Checked the I.D. to a certified gage ring and get an accurate reading. Then we check the O.D to a gage pin where the measurement will be off be .006 to .008 microns. If we adjust the size of probe ball dia. we can the I.D. to measure correct but then O.D. is not. The Master Probe is qualifying at 3.9994. Also checked Master probe and Master sphere for cracks or blemishes.

Any help would be appreciated, thanks in Advanced
2nd what Richard said on the class of gauges and their prescribed accuracy.
What size is the pin gauge and how are you holding it?
Is it (pin gauge) big or long enough to measure in two different locations, towards the bottom and at the top if measuring in the Z axis. The reason I ask is that if you get two different sizes at different lengths with a significant difference in size (significant pending the accuracy needed) you're probably seeing deflection. The method the pin is being held and how much of the surface is being held is critical because if it's not secure enough, the part will deflect. I've seen that happen many times.
Both Ring Gage and pin gage are Calibrated/Certified By A2 measuring house to an accuracy of .00001. I am creating base alignments on each gage to ensure Perpendicular contact. Nothing making sense.... for the record this Contura has been a LEMON ,since the day we bought it brand new from Zeiss.
.

Elaborate on "LEMON." I'm curious. 🧑‍💻

Also, what is the weight and length of the probe you are using? If you have an XXT sensor, they tend to not handle long, heavy probes.

Here's an example of a cylindrical square I probed on a Contura/XXT Sensor with a 125mm long / Ø8mm Thermofit (carbon fiber) stylus. There was no taper in the square. It's all probing error due to length/weight.



.

CAD_Evaluation_cylindrical_square.png

Please sign in to view this quote.

The XXT works like a fulcrum. The shorter the fulcrum the more force that is required to move it. The longer the fulcrum the less force that is required to move it.
659_2b47bc199df790bbf0c3d6eb2f42a5de.pdf
I use this to my advantage when measuring non-rigid parts.
×
×