Hello,
While inspecting the pink part that you may see in background in the attached ppt I noticed there wer some discrepancies in the position and line profiles vs the supplier CMM report (done in a Hexagon) so I decided to test the accuracy of the O-inspect with a .6303 pin gage and for my dismay the position of one end cylinder was out respect to the other(see report attached) end's in spite both cylinders are the same and in case someone ask the bright question the pin is 2x and was recently calibrated.

0.6303 PIN GAGE MEASUR.pptx

Link to comment
Share on other sites

Just a thought, what is cylinder 1's depth? Should be at least 3x the Ø. Try measuring cylinder 2 as a circle.
Also, to rule out the machine as the problem, try moving the part somewhere else on the O-inspect and
re-measure it. If it is the machine, then it might be a specific location that is out of spec.
Link to comment
Share on other sites