What would it take to have a CMM run a probe qualification program without the help of a human? Is there some way to set a re-occurring timer/schedule that would/could make the CMM run qualification program? We have 3 Duramax's that the sphere is bolted in place and is only moved during machine calibration. I have programs in Autorun that run through each rack and qualify each stylus.

Is this possible to do?

Asking for a friend!!!! Ha!
Link to comment
https://qualityforum.zeiss.com/topic/11866-automatic-probe-qualification/
Share on other sites

Hello,

I agree with Gilbert, it is certainly possible.
However, I would not recommend an automatic qualification of the probes.
Before performing this task, it is essential to thoroughly clean all probes and check their condition. The smallest particle/fiber will greatly distort the results.
I think that the qualification should be started manually by experienced staff and under optimal conditions.
You can of course write a program containing a PCM waiting routine, but you would have to start it some time before calibration time and couldn't run anything else inbetween.
It's not possible to have a program load and start automatically without some additional external control instance being present (like a Zeiss software or option, whatever that would be). PCM only runs from within an already running program, even if you use external PCM files.
×
×