A sister plant is attempting to use an older Vast XT, but using a competitors software with the help of EasyCMM. I've been asked to assist, but I've never heard of this EasyCMM before. I've read the manual and have a good idea of what I need to do. However, I'm unaware of how to manually remove the currently active probe with out clicking on the "manual stylus system change" button in Calypso. Is there a way to do this? Whether in Easy CMM, or *cough* PC-DMIS, or without any software?

Thanks in advance.
Link to comment
https://qualityforum.zeiss.com/topic/11964-easycmm-with-vast-xt/
Share on other sites

Oh, and I remember trying to use this key code to try this when I was using an XXT, and it really releases something, but it’s not the probe, but the whole sensor, it’s the whole XXT that releases. There is a safety pin that should prevent the XXT from falling, but it’s made of plastic and could fail, so never use this key code on an XXT without your hand firmly gripping the sensor, or it will be a very expensive error.
Link to comment
https://qualityforum.zeiss.com/topic/11964-easycmm-with-vast-xt/#findComment-59586
Share on other sites

No, usually you don’t have to remove the sensor. But the key combination releases the https://mediaserver.probes.zeiss.com/images/cimage_86f97a985449bc4cb573e850bbd48a52.ashx, which could drop it on the granite table (except there’s a safety pin catching it most of the times) along with the XXT, which is really pricey.
Link to comment
https://qualityforum.zeiss.com/topic/11964-easycmm-with-vast-xt/#findComment-59590