Master Probe wont Qualify the Calibration Ball
You are selecting Reference Sphere Position with MasterProbe attached correct?
I would use qualify passive when referencing sphere position and see if the error goes away. Also make sure the R value looks correct 2.5 mm. Another thing to try is to Reference Sphere Position outside of a program and see if the issue goes away.
  • 2 years later...
We had this problem in July Error 1105 master probe would not qualify the calibration sphere... using Calypso 2020.
As a temporary fix, we were able to use calypso 2019 for a few weeks and run qualification of probe styli on 2019.
However, we were unable to use all our programs that were written using calypso 2020 on calypso 2019... this helped for a few weeks in limited fashion since 95% of our programs we use are written in 2020.
This was not resolved until the 3rd Zeiss tech from Minnesota was able to walk us through resetting the settings for the system/ master probe in the software. the first two were unable to help but the last one did know what to do.... amazing complex settings but not impossible to navigate.... even the trainer was baffled.
Goes to show you that you need to know to keep asking for help from the Technicians, this saved us an expensive in-person visit from Zeiss....
cleaned off computer and moved a lot of old, duplicate files to free up space
removed and reinstalled Calypso 2020
updated the software with service pack
service call from a third Zeiss tech was able to access our settings and reset the VAST head settings..
not certain this was accurate recall of events, just shout out to the technician who was able to set us up again..
have been running well since late July?
sometimes the computer will respond in German on Calypso software, when I remove a section of a program, it does.
shut down computer and CMM and reboot and seems to work.