Recommended Posts

  • 2 weeks later...
Here's a good youtube video on roundness but at 12:43 in the video, where he discusses the minimum zone, he does not clearly state that the 2 circles are concentric but you can clearly see that each circle is only making contact with extreme lowest and highest single point, not the Max Inscribed and Min Circumscribed.

https://youtu.be/NjbvOTUSqdI

Update: I stand corrected. I was wrong about by observation mentioned above. After watching the video again, the minimum zone shows the 2 circles making contact in 3 places. And the resultant roundness was the deviation between the min circumscribed and max inscribed circles. I still am of the opinion this is incorrect. Using that logic, how do I evaluate the 2 circles below?

Screenshot 2021-07-29 055756.jpg