First up all 30mm per second is not high speed, I have been working with contura G2 more than 3 years and I did not observe any jerking as you mentioned. The only way to control acceleration is controlling the speed but you are already running in minimum level speed I think.
Link to comment
https://qualityforum.zeiss.com/topic/12510-bridge-acceleration/#findComment-62658
Share on other sites

I would also be interested in this, but for a different reason. Our micura when running more than a speed of 60 to 65, will sometimes when accellerating in the Y direction (so the bridge has to move) will use enough air that the inlet pressure drops too low and the drives kick off, requiring us to re-enable the drives, switch the stoplight to green, and restart whatever routine or program we were running. If we could sacrifice a little accelleration to prevent this, it might be worth it.
Link to comment
https://qualityforum.zeiss.com/topic/12510-bridge-acceleration/#findComment-62660
Share on other sites

Please sign in to view this quote.

If you are working in tight areas you can limit navigation within the feature. You can then speed up the program navigation and it will run fast between features and probe changes, but slowly inside small feature strategy.

Measurement Plan Editor Features> Speed>selected features
Link to comment
https://qualityforum.zeiss.com/topic/12510-bridge-acceleration/#findComment-62662
Share on other sites

Maybe the speed mentioned above was a typo? I think it should be more like 300 mm/s, right?

Indeed there is no way to set the acceleration in Calypso. I never understood this, since in all predecessors (UMESS) this was a a standard setting.
But there is another option: On C99 controllers (at least), acceleration can be set per axis in a parameter in the controller. Unforntunately I don't remember the parameter name right now. When I have some spare time, I'll try to look it up. It might even be possible to set it via the undocumented cmmInquire() PCM function (but I never tried).
Apart from that, only a service tech can change that parameter with his servicing software.