[To...] Posted March 19, 2021 Share Posted March 19, 2021 On RDS-XXT and get the warning below after attempting Reference Sphere Position. Any thoughts? SDO Error -1117 No Convergence 1 Link to comment https://qualityforum.zeiss.com/topic/13294-trouble-with-ref-sphere-position/ Share on other sites More sharing options...
[Er...] Posted March 19, 2021 Share Posted March 19, 2021 Try the suggestions in this post. Good luck. viewtopic.php?f=11&t=1849 I edited it to add the suggestions. t could be a number of things. Follow this list: 1. Restart the computer with Calypso and power the CMM controller off for 30 seconds then power back on. 2. Make sure everything is clean: MasterProbe sphere, reference sphere, stylus adapter plate, probehead adapter plate mounting surface, and RDS adapter plate and mating surface (if using an RDS). 3. Make sure everything is tight: no loose spheres on MasterProbe or reference sphere, reference sphere clamped down solid. Hopefully this helps. I have seen this in the past with a Vast XXT with RDS but have not experienced it with any other sensors. Link to comment https://qualityforum.zeiss.com/topic/13294-trouble-with-ref-sphere-position/#findComment-67602 Share on other sites More sharing options...