[ke...] Posted March 1, 2021 Share Posted March 1, 2021 Running a Probe Qualification program, I've run this one before. When the sphere is in the #4 position (see photo) the Master Probe is coming down in the #1 position. I've deleted the Master Probe information from this part of the program, entered it again, and also copy and pasted it from another program but it still comes down in the #1 position missing the sphere. Link to comment https://qualityforum.zeiss.com/topic/13390-master-probe-missing-sphere-during-probe-qualification/ Share on other sites More sharing options...
[No...] Posted March 2, 2021 Share Posted March 2, 2021 Have you tried to manually requalify the sphere at the #4 position? Calypso remembers the last manual position of a sphere. Link to comment https://qualityforum.zeiss.com/topic/13390-master-probe-missing-sphere-during-probe-qualification/#findComment-68210 Share on other sites More sharing options...
[ke...] Posted March 2, 2021 Author Share Posted March 2, 2021 Dang, it. Thanks. Link to comment https://qualityforum.zeiss.com/topic/13390-master-probe-missing-sphere-during-probe-qualification/#findComment-68212 Share on other sites More sharing options...