[Ar...] Posted July 26, 2019 Share Posted July 26, 2019 Has anyone experience error codes: -1012 "Error in the SDO durin the calculation for the qualification for linear sensors and during the definition of the start value for non-linear sensors" or error code 1215 - "The sigma of the sphere measurement exceeds the default value defined for this machine. Please check the sigma value of the stylus to see if it meets your requirements." I got these errors while trying to re-qualify the MasterProbe... I appreciate the help on what I need to resolve this... Link to comment Share on other sites More sharing options...
[Ja...] Posted August 6, 2019 Share Posted August 6, 2019 Armando, It could be a number of things. Follow this list: 1. Restart the computer with Calypso and power the CMM controller off for 30 seconds then power back on. 2. Make sure everything is clean: MasterProbe sphere, reference sphere, stylus adapter plate, probehead adapter plate mounting surface, and RDS adapter plate and mating surface (if using an RDS). 3. Make sure everything is tight: no loose spheres on MasterProbe or reference sphere, reference sphere clamped down solid. Hopefully this helps. I have seen this in the past with a Vast XXT with RDS but have not experienced it with any other sensors. Link to comment Share on other sites More sharing options...
[Ma...] Posted February 3, 2021 Share Posted February 3, 2021 I am not the OP of this thread but thank you. This reply was very helpful to us. Somehow, the Vast XXT holder was loose from the RDS. Link to comment Share on other sites More sharing options...