[Ar...] Posted July 26, 2019 Share Posted July 26, 2019 Has anyone experience error codes: -1012 "Error in the SDO durin the calculation for the qualification for linear sensors and during the definition of the start value for non-linear sensors" or error code 1215 - "The sigma of the sphere measurement exceeds the default value defined for this machine. Please check the sigma value of the stylus to see if it meets your requirements." I got these errors while trying to re-qualify the MasterProbe... I appreciate the help on what I need to resolve this... Link to comment https://qualityforum.zeiss.com/topic/13608-error-code-1012-1215-during-masterprobe-qualification/ Share on other sites More sharing options...
[Ja...] Posted August 6, 2019 Share Posted August 6, 2019 Armando, It could be a number of things. Follow this list: 1. Restart the computer with Calypso and power the CMM controller off for 30 seconds then power back on. 2. Make sure everything is clean: MasterProbe sphere, reference sphere, stylus adapter plate, probehead adapter plate mounting surface, and RDS adapter plate and mating surface (if using an RDS). 3. Make sure everything is tight: no loose spheres on MasterProbe or reference sphere, reference sphere clamped down solid. Hopefully this helps. I have seen this in the past with a Vast XXT with RDS but have not experienced it with any other sensors. Link to comment https://qualityforum.zeiss.com/topic/13608-error-code-1012-1215-during-masterprobe-qualification/#findComment-69784 Share on other sites More sharing options...
[Ma...] Posted February 3, 2021 Share Posted February 3, 2021 I am not the OP of this thread but thank you. This reply was very helpful to us. Somehow, the Vast XXT holder was loose from the RDS. Link to comment https://qualityforum.zeiss.com/topic/13608-error-code-1012-1215-during-masterprobe-qualification/#findComment-69786 Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in