Hello all.

Our expiration date on the CMM's battery backup is coming up. At first when I saw this I thought this cant be true. I was wrong. The CMM is connected to a battery backup (APC Model: Back-UPS 900) which I think is too small for a CMM even if its just idling.

Have any of you needed a battery backup? And do any of you have a story about the power shutting off in the middle of a run?

I'm curious but not curious enough to test it on our CMM.


Thanks in advanced!
Link to comment
https://qualityforum.zeiss.com/topic/14656-battery-backup-for-cmm/
Share on other sites

We have some UPS but i don't know details. But once we died in middle of measuring. With some effort we succesfully continued after power on. I don't know how big should be battery to backup and stand for power loss for a few sec. I saw only 2 or 3 blackouts for more than 30 mins in 13 years in factory.
Link to comment
https://qualityforum.zeiss.com/topic/14656-battery-backup-for-cmm/#findComment-77172
Share on other sites

That UPS is enough to run a smaller machine for a brief time period. If I take the Contura's spec sheet rating for typical draw at 200 watts, APC's site gives an estimate of 13 minutes of battery time. A Duramax would run for 20. Depending on what CMM you're running, and whether or not you also have the controlling PC hooked up to the battery (you probably should), consider upsizing.

Regardless, I wouldn't *run* a CMM on battery. If there's a power outage that I don't expect to be resolved for a while, the UPS is there so I have enough time to cleanly shut down my equipment.